Investigations

1st investigations to order

transthoracic echo with colour Doppler and contrast injection

Test
Result
Test

Performed as a screening test when clinical suspicion of condition is low.

Intravenous access is obtained, typically in an antebrachial vein or femoral vein.

Two 10 mL syringes are attached to a 3-way stopcock. About 9.5 mL of normal saline and 0.5 mL of air are drawn into a syringe and then rapidly agitated via the stopcock and the other syringe just prior to intravenous injection.

Echo images are continuously recorded over several cardiac cycles and the sequence is repeated several times with provocative manoeuvres.

If upper extremity injection fails to show crossover despite a high suspicion of PFO, lower extremity injection may be used as this can be more sensitive.[20]

Result

colour flow at the interatrial septum and crossover of microbubbles from right atrium to left atrium after intravenous injection of agitated saline

trans-oesophageal echo with colour Doppler and contrast injection

Test
Result
Test

Performed in cases where a definitive diagnosis is sought; the transthoracic echo is either equivocal or negative; there is uncertainty as to whether the patient has atrial septal defect or PFO; and diagnosis is likely to alter management. [Figure caption and citation for the preceding image starts]: Trans-oesophageal echo showing a 2D image of the interatrial septum with location of the PFOFrom the collection of Kul Aggarwal, MD [Citation ends].com.bmj.content.model.Caption@5ecb14e2[Figure caption and citation for the preceding image starts]: Trans-oesophageal echo with intravenous injection of agitated saline showing contrast crossing through PFO into left atriumFrom the collection of Kul Aggarwal, MD [Citation ends].com.bmj.content.model.Caption@418da991[Figure caption and citation for the preceding image starts]: Trans-oesophageal echo with colour flow Doppler showing presence of PFOFrom the collection of Kul Aggarwal, MD [Citation ends].com.bmj.content.model.Caption@280c9f4a

Result

visualisation of defect, colour flow at the interatrial septum and crossover of microbubbles from right atrium to left atrium after intravenous injection of agitated saline

Investigations to consider

intracardiac echo

Test
Result
Test

Not routinely used as a diagnostic procedure for PFO because of its high cost.

Performed in conjunction with percutaneous closure of PFO.[21]

Advantage is elimination of need for deeper sedation and discomfort associated with a trans-oesophageal echo.[Figure caption and citation for the preceding image starts]: Intracardiac echo showing a catheter placed across the PFO into the left atriumFrom the collection of Kul Aggarwal, MD [Citation ends].com.bmj.content.model.Caption@2e00b08

Result

visualisation of PFO

ultrasound lower extremity

Test
Result
Test

Useful if suspected paradoxical embolism.

Result

deep venous thrombosis

MRI lower extremity

Test
Result
Test

Useful if suspected paradoxical embolism.

Result

deep venous thrombosis

tilt-table trans-oesophageal echo

Test
Result
Test

Diagnosis of platypnoea-orthodeoxia syndrome is best made by tilt-table trans-oesophageal echo. A diagnosis of PFO is established if an inserted TOE probe shows right-to-left shunting across the PFO resulting in hypoxaemia in an upright position, and if there is a resolution of shunting and hypoxaemia when the patient is in the recumbent position.

Result

right-to-left shunting and hypoxaemia in the upright position; resolution in the recumbent position

Emerging tests

transcranial Doppler

Test
Result
Test

Modality not in common clinical use.

May be used in centres with particular expertise and where echo images are suboptimal.

Result

microbubbles in the cranial circulation immediately after intravenous saline